LA502T Tube Optical Compressor / Re: test procedure - having trouble
« Last post by doktar on October 08, 2026, 07:46:56 PM »After repeating, Im now getting about 23V max in step 5. So good there
Step 6 - with peak at 0 - there is 0v AC/DC on TP5.
With peak adjusted to 10.6V - i get only -3 LED illuminated, but can adjust because all trimmers are down.
With peak at 1.5, All LEDS are illuminated. Still no adjustment.
I'm closer now, but seems like slightly higher GR meter readings. The -20 illuminates around 2.2 V
Step 6 - with peak at 0 - there is 0v AC/DC on TP5.
With peak adjusted to 10.6V - i get only -3 LED illuminated, but can adjust because all trimmers are down.
With peak at 1.5, All LEDS are illuminated. Still no adjustment.
I'm closer now, but seems like slightly higher GR meter readings. The -20 illuminates around 2.2 V